Topics Map > Standard Operating Procedures (SOP)
Fluid Bag Use and Storage SOP
Printable PDF: Fluid Bag Use and Storage SOP
Purpose/ Applicability:
To provide clear guidelines for the warming, storage, and use of sterile fluid bags for parenteral or external administration, to maintain sterility and product integrity.
Scope:
This SOP applies to all sterile fluid bags used for parenteral or external use, including but not limited to Sodium Chloride 0.9%, Lactated Ringer’s Solution, Normosol-R, Plasma-Lyte, Mannitol 20% and Hypertonic Saline 7.2%.
Certain products have special handling or warming restrictions, which are outlined in the Important Exceptions section of this SOP.
General Safety Requirement:
- All sterile fluid bags and bottles must be inspected for clarity, intact seals, and absence of visible contamination before each use.
- Any container showing cloudiness, particulates, discoloration, leakage, or other signs of compromise must be discarded immediately, regardless of the product type, storage conditions, or allowable use period.
- AAHA Accreditation Requirement (John A. Coyne South Clinic):
A new container of fluids and a new administration set must be used for each patient, regardless of the route of administration, in compliance with AAHA Standards.
Procedure:
Labeling Requirements
To ensure accurate tracking of discard times and avoid relying solely on manufacturer expiration dates:
- All fluid bags/bottles must be labeled with the date and time when:
- Removed from protective overwrap (opened)
- Placed into a fluid warmer
- Fluid Solutions (bag or bottle) – OPENED
Room Temp/Refrigerated
-
- Opened fluid bags (i.e., removed from the protective overwrap) may be stored for up to 24 hours at room temperature or under refrigeration.
- External use only (e.g., wound irrigation):
- With a line attached, may be used for up to 30 days or until the expiration date, whichever comes first.
- Without a line, withdrawn via a syringe/needle, may be used for up to 7 days.
- Must be clearly labeled: “For External Use Only”
Warmed Fluids – OPENED (Removed from Warmer)
-
- Opened fluid bags removed from the warmer and stored at room temperature must be used within 24 hours.
- Best practice: Discard at the end of the business day.
- Opened fluid bags removed from the warmer and stored at room temperature must be used within 24 hours.
- Important Exceptions
-
- Dextrose-containing solutions and Hetastarch: Do not warm under any circumstances.
- Mannitol 20% (VetOne, 100ml vial):
- If crystals form, warm in hot water or autoclave for 15 minutes.
- Storage recommended at 15°C - 30°C.
- Hypertonic Saline 7.2% PF (VetOne, 1L bottle):
- Not recommended for long-term storage in a fluid warmer.
- If necessary, the fluid warmer must be set to ~37°C; do not exceed 40°C.
- In all cases, the bag or bottle must be maintained in aseptic conditions and remain visibly uncontaminated.
- Fluid Solutions (bag or bottle) – UNOPENED
-
- Warmed unopened solutions may not be rewarmed or returned to regular room temperature storage beyond 24 hours.
|
Product Type |
Max warming temperature |
Max warming period UNOPENED |
Use After Removal UNOPENED |
|
Irrigation solution - BAG |
≤ 104°F (40°C) |
14 days |
Discard within 24 hours |
|
Irrigation Pour BOTTLE |
≤ 122°F (50°C) |
14 days |
Discard within 24 hours |
|
Parenteral Use |
≤ 104°F (40°C) |
14 days*
|
Discard within 24 hours |
|
*Only if ≥ 3 months remain before product expiration |
|||
Definition(s):
Aseptic Technique: A method used to prevent contamination by pathogens during the handling or preparation of sterile products.
External Use: Use of fluids for non-parenteral purposes (e.g., wound flushing or lavage), not intended for injection.
Opened Bag: A fluid bag from which the protective overwrap has been removed or has been accessed via spike, needle, or other port.
Parenteral Use: Administration of fluids via injection routes (e.g., intravenous, subcutaneous).
Unopened Bag: A fluid bag that remains sealed in its original protective overwrap and has not been accessed.
Warmed Fluids: Sterile fluid bags or bottles that have been placed in an approved fluid warmer to increase temperature before administration